Quarter Bridge Strain Task shows error 200077

Asked By 0 points N/A Posted on -
qa-featured

Hello,

When I try to configure Quarter Bridge Strain Task I get the error 200077. Why I got such error?

Primary Software: Driver Software>>NI-DAQ

Primary Software Version: 4.6.1

Primary Software Fixed Version: N/A

Secondary Software: N/A

Hardware: Multifunction DAQ (MIO)>>SC Series>>PXIe-4330

 

Here is my problem;

I tried to adjust the nominal gage resistance in my PXIe-4330 in quarter bridge configuration. I get mentioned error at DAQmx Start Task.vi when trying to configure a strain task by Cont Acq Strain Samples.vi.

I have given an snapshot below.

What is the reason for this?

Give me a solution please.

Thank you.

Error -200077 occurred at DAQmx Start Task.vi:2

Possible reason(s):

Requested value is not a supported value for this property. The property value may be invalid because it conflicts with another property

Property: AI.Bridge.NomResistance

Requested Value: 250.0

You Can Select : 120.0, 350.0, 1.0e3

Channel Name: SCExpress/ai0

SHARE
Answered By 0 points N/A #177503

Quarter Bridge Strain Task shows error 200077

qa-featured

Hi Tresa,

That error message appeared because you were trying to specify a gage resistance which is not supported by the module of the software. You must specify the Nominal Gage Resistance value into 250.00 if you are trying to configure using Quarter Bridge. You must remember that  the PXIe-4330 supports gage resistance of 120, 350, and 1000 Ohm. See the illustration below: